SETUP VARIABLE
PHYS_RANGE2
CONC_RANGE1
CONC_RANGE2
11
CO2_RANGE
10
O2_RANGE
SAMP_FLOW_SET
SAMP_FLOW_SLOPE
3
VAC_SAMP_RATIO
SAMP_PRESS_SET
3
VAC_PRESS_SET
BOX_SET
PMT_SET
NUMERIC
DEFAULT
UNITS
VALUE
PPM
22,
3
5500
Conc
500,
3
5000
Conc
500,
3
5000
%
15
%
100
cc/m
700,
1+9
250
Warnings:
350–1200,
1+9
175–325
1
0.53
"Hg
29.92
Warnings:
15–35
"Hg
6
Warnings:
3–10
ºC
30
Warnings:
8–50
ºC
7,
12
15
Warnings:
2–12,
12
2–20
05036 Rev D
APPENDIX A - Version Specific Software Documentation
VALUE RANGE
0.1–2500,
High pre-amp range.
3
5–10000
0.1–50000
D/A concentration range 1.
0.1–50000
D/A concentration range 2.
0.1–500
CO
2
0.1–500
O
concentration range.
2
0–6000
Sample flow set point for flow
calculation and warning limits.
0.5–1.5
Sample flow slope correction factor
(adjusted flow = measured flow x
slope).
0.1–2
Maximum vacuum pressure / sample
pressure ratio for valid sample flow
calculation.
0–100
Sample pressure set point for pressure
compensation and warning limits.
0–100
Vacuum pressure set point for
pressure compensation and warning
limits.
5–60
Box temperature warning limits. Set
point is not used.
0–40
PMT temperature set point and
warning limits.
DESCRIPTION
concentration range.
A-15