4.2. SAMPLE MODE

4.2.1. TEST FUNCTIONS

80
The Mode field of the front panel display indicates to the user which operating mode the
unit is currently running.
Besides SAMPLE and SETUP, other modes the analyzer can be operated in are:
Table 4-1: Analyzer Operating Modes
One of the analyzer's diagnostic modes is active (refer to Section 4.8).
Unit is performing LOW SPAN (midpoint) calibration initiated automatically by the analyzer's
AUTOCAL feature
Unit is performing LOW SPAN (midpoint) calibration initiated remotely through the COM ports or
digital control inputs.
This is the basic calibration mode of the instrument and is activated by pressing the CAL button.
Sampling normally, flashing text indicates adaptive filter is on.
Indicates that unit is in SAMPLE mode and AUTOCAL feature is activated.
SETUP mode is being used to configure the analyzer. The gas measurement will continue during
this process.
Unit is performing SPAN calibration initiated automatically by the analyzer's AUTOCAL feature
Unit is performing SPAN calibration initiated manually by the user.
Unit is performing SPAN calibration initiated remotely through the COM ports or digital control
inputs.
Unit is performing ZERO calibration procedure initiated automatically by the AUTOCAL feature
Unit is performing ZERO calibration procedure initiated manually by the user.
Unit is performing ZERO calibration procedure initiated remotely through the COM ports or digital
control inputs.
Finally, the various CAL modes are for calibrating the analyzer. This mode is described
separately in Section 6.
This is the analyzer's standard operating mode. In this mode, the instrument is analyzing
SO
and calculating concentrations.
2
A series of test functions is available at the front panel while the analyzer is in
SAMPLE mode. These parameters provide information about the present operating
status of the instrument and are useful during troubleshooting (refer to Section 10.1.2).
They can also be recorded in one of the DAS channels (refer to Section 5.2) for data
analysis. To view the test functions, press one of the <TST TST> buttons repeatedly in
either direction.