Troubleshooting & Repair
10.7.2.3. CLEANING THE PMT LENS AND PMT FILTER
IMPORTANT
Figure 10-9:
288
The sample chamber should only be opened or cleaned on instructions from
the Teledyne ML Customer Service Department.
Be careful not to leave thumbprints on the interior of the sample chamber. The
various oils that make up fingerprints fluoresce brightly under UV light and will
significantly affect the accuracy of the analyzer's SO
To clean the PMT Lens and filter:
1. Remove the sensor module as described in Section 10.7.2.1.
Hex Screw Between Lens Housing and Sample chamber
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
06807A(DCNxxxx)