Dual Range Mode (Dual); Figure 4-12: Setup Rnge - Dual Range Mode - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML - T100 UV Fluorescence SO2 Analyzer

4.6.5. DUAL RANGE MODE (DUAL)

range1: The range setting for the A1 output.
range2: The range setting for the A2 output.
SAMPLE
RANGE = 500.000 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
PRIMARY SETUP MENU
SETUP X.X
CFG DAS
RNGE
PASS CLK MORE
SETUP X.X
RANGE CONTROL MENU
MODE
SET UNIT
SETUP X.X
RANGE MODE:
SNGL
DUAL
AUTO
.
IMPORTANT
Selecting Dual Range mode allows the A1 and A2 outputs to be configured with
different reporting ranges. The analyzer software calls these two ranges low and high.
The Low range setting corresponds with the analog output labeled A1 on the rear panel
of the instrument. The high range setting corresponds with the A2 output. While the
software names these two ranges low and high, they do not have to be configured that
way. For example: the low range can be set for a span of 0-150 ppb while the high range
is set for 0-50 ppb.
In DUAL range mode the RANGE test function displayed on the front panel will be
replaced by two separate functions:
To set the ranges press following control button sequence
SO2 =XXX.X
SETUP
ENTR
EXIT
EXIT
EXIT
SNGL
ENTR EXIT
SETUP RNGE – Dual Range Mode
Figure 4-12:
In DUAL range mode the LOW and HIGH ranges have separate slopes and offsets for
computing SO
concentration. The two ranges must be independently calibrated.
2
SETUP X.X
RANGE MODE:
DUAL
SNGL DUAL AUTO
SETUP X.X
RANGE CONTROL MENU
MODE
SET
UNIT
SETUP X.X
LOW
RANGE:
500.0 Conc
0
0
1
0
0
.0
SETUP X.X
HIGH RANGE: 500.0 Conc
0
0
5
0
0
.0
SETUP X.X
RANGE CONTROL MENU
MODE SET UNIT
Operating Instructions
ENTR
EXIT
EXIT
Toggle the
ENTR
EXIT
numeral buttons
to set the upper
limit of each
range.
ENTR
EXIT
EXIT Returns
to the Main
EXIT
SAMPLE Display
95