Analog I/O Configuration; Table 4-15: Diag - Analog I/O Functions; Table 4-16: Analog Output Voltage Ranges - Teledyne T360 Operation Manual

Carbon dioxide analyzer
Table of Contents

Advertisement

Teledyne API Model T360/T360M Operation Manual

4.13.4. Analog I/O Configuration

Table 4-15: DIAG - Analog I/O Functions

Sub Menu
AOUTS CALIBRATED:
CONC_OUT_1
CONC_OUT_2
TEST OUTPUT
AIN CALIBRATED
XIN1
.
.
.
XIN8
07272B DCN6552
The analog I/O functions that are available in the T360 are:
Shows the status of the analog output calibration (YES/NO) and initiates a calibration of all
analog output channels.
Sets the basic electronic configuration of the A1 analog output (CO
Range: Selects the signal type (voltage or current loop) and full scale level of the output.
REC_OFS: Allows setting a voltage offset (not available when RANGE is set to CURRent loop.
Auto_CAL: Performs the same calibration as AOUT CALIBRATED, but on this one channel only.
NOTE: Any change to RANGE or REC_OFS requires recalibration of this output.
Same as for CONC_OUT_1 but for analog channel 2 (CO
Same as for CONC_OUT_1 but for analog channel 4 (TEST)
Shows the calibration status (YES/NO) and initiates a calibration of the analog to digital
converter circuit on the motherboard.
For each of 8 external analog input channels, shows the gain, offset, engineering units, and
whether the channel is to show up as a Test function.
To configure the analyzer's three analog outputs, set the electronic signal type of
each channel and calibrate the outputs. This consists of:
Selecting an output type (voltage or current, if an optional current output
driver has been installed) and the signal level that matches the input
requirements of the recording device attached to the channel, see Section
4.13.4.1.
Calibrating the output channel. This can be done automatically or manually
for each channel, see Section 4.13.4.2 and 4.13.4.3.
Adding a bipolar recorder offset to the signal, if required (Section 4.13.4.2).
In its standard configuration, the analyzer's outputs can be set for the following
DC voltages. Each range is usable from -5% to + 5% of the nominal range.

Table 4-16: Analog Output Voltage Ranges

RANGE
MINIMUM OUTPUT
0-0.1 V
0-1 V
0-5 V
0-10 V
The default offset for all ranges is 0 VDC.
Function
MAXIMUM OUTPUT
-5 mV
-0.05 V
-0.25 V
-0.5 V
Operating Instructions
). There are three options:
2
)
2
+105 mV
+1.05 V
+5.25 V
+10.5 V
135