Table 7-1: Activity Matrix For Calibration Equipment & Supplies - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

EPA Protocol Calibration
Table 7-1: Activity Matrix for Calibration Equipment & Supplies
EQUIPMENT &
SUPPLIES
Recorder
Sample line and
manifold
Calibration equipment
Working standard SO
cylinder gas or SO
2
permeation tube
Zero air
Record form
Audit equipment
198
Calibrations should be carried out at the field monitoring site. The Analyzer should be in
operation for at least several hours (preferably overnight) before calibration so that it is
fully warmed up and its operation has stabilized. During the calibration, the T100 should
be in the CAL mode, and therefore sample the test atmosphere through all components
used during normal ambient sampling and through as much of the ambient air inlet
system as is practicable. If the instrument will be used in more than one range (i.e.
DUAL OR AUTO ranges), it should be calibrated separately on each applicable range.
Calibration documentation should be maintained with each analyzer and also in a central
backup file.
ACCEPTANCE LIMITS
Compatible with output
signal of analyzer; min.
chart width of 150 mm (6 in)
is recommended
Constructed of PTFE or
glass
Meets guidelines of
reference 1 and Section
2.3.2 (Q.A. Handbook)
Traceable to NIST-SRM
2
meets limits in traceability
protocol for accuracy and
stability (refer to Section
2.0.7, Q.A. Handbook)
Clean dry ambient air, free
of contaminants that cause
detectable response with
the SO
analyzer.
2
Develop standard forms
Must not be the same as
used for calibration
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
FREQUENCY AND METHOD
OF MEASUREMENT
Check upon receipt
Check upon receipt
Refer to Section 2.3.9 (Q.A.
Handbook)
Analyzed against NIST-SRM;
refer to protocol in Section
2.0.7, Q.A. Handbook
Refer to Section 2.9.2 (Q.A.
Handbook)
N/A
System must be checked out
against known standards
06807A(DCNxxxx)
ACTION IF
REQUIREMENTS ARE NOT
MET
Return equipment to supplier
Return equipment to supplier
Return equipment/ supplies
to supplier or take corrective
action
Obtain new working
standard and check for
traceability
Obtain air from another
source or regenerate.
Revise forms as appropriate
Locate problem and correct
or return to supplier