When editing the data channels, the top line of the display indicates some of the
configuration parameters. For example, the display line:
0) CONC: ATIMER, 4, 800
represents to the following configuration:
CHANNEL NUMBER.: 0
NAME: CONC
TRIGGER EVENT: ATIMER
PARAMETERS: Four parameters are included in this channel
EVENT: This channel is set up to record 800 data points.
To edit the name of a data channel, refer to Figure 5-11, then press:
SETUP X.X
NAME:CONC1
<SET
SET>
EDIT
SETUP X.X
NAME:CONC
C
O
N
C
Press each key repeatedly to cycle through the
available character set:
0-9, A-Z, space ' ~ !  # $ % ^ & * ( ) - _ = +[ ]
{ } < >\ | ; : , . / ?
Figure 5-12:
Remote Operation & Advanced Features
PRINT
EXIT
1
-
ENTR
EXIT
DAS – Editing Data Channel Name
ENTR accepts the new string
and returns to the previous
menu.
EXIT ignores the new string
and returns to the previous
menu.