Teledyne T100 Instruction Manual page 340

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML - T100 UV Fluorescence SO2 Analyzer
Table A-3:
TEST FUNCTION
RANGE
RANGE=500.0 PPB
SO2 RNG=500.0 PPB
RANGE1
RANGE1=500.0 PPB
SO2 RN1=500.0 PPB
RANGE2
RANGE2=500.0 PPB
SO2 RN2=500.0 PPB
CO2RANGE
CO2 RNG=100 PCT
O2RANGE
O2 RNG=100 PCT
STABILITY
STABIL=0.0 PPB
SO2 STB=0.0 PPB
O2 STB=0.0 PCT
CO2 STB=0.0 PCT
6
STABILITY2
STABIL2=0.0 PPB
SO2 STB2=0.0 PPB
O2 STB2=0.0 PCT
CO2 STB2=0.0 PCT
2
RESPONSE
RSP=1.11(0.00) SEC
5
VACUUM
VAC=9.1 IN-HG-A
SAMPPRESS
PRES=29.9 IN-HG-A
SAMPFLOW
SAMP FL=700 CC/M
PMTDET
PMT=762.5 MV
NORMPMTDET
NORM PMT=742.9 MV
UVDET
UV LAMP=3457.6 MV
6
STABILITYUV
UV STB=5.607 MV
LAMPRATIO
LAMP RATIO=100.0 %
STRAYLIGHT
STR. LGT=0.1 PPB
DARKPMT
DRK PMT=19.6 MV
DARKLAMP
DRK LMP=42.4 MV
SLOPE
SLOPE=1.061
OFFSET
OFFSET=250.0 MV
11
CO2SLOPE
CO2 SLOPE=1.0000
11
CO2OFFSET
CO2 OFFSET=0.00 %
10
O2SLOPE
O2 SLOPE=0.980
10
O2OFFSET
O2 OFFSET=1.79 %
HVPS
HVPS=650 VOLTS
RCELLDUTY
RCELL ON=0.00 SEC
RCELLTEMP
RCELL TEMP=52.1 C
10
O2CELLTEMP
O2 CELL TEMP=50.2 C
BOXTEMP
BOX TEMP=35.5 C
Test Functions, Software Version G.4
MESSAGE TEXT
Test Measurements
3
3, 10, 11
3
3, 10, 11
3
3, 10, 11
11
10
3
3, 10
10
11
3
3, 10
10
11
05036 Rev D
APPENDIX A - Version Specific Software Documentation
DESCRIPTION
D/A range in single or auto-range modes.
D/A #1 range in independent range mode.
D/A #2 range in independent range mode.
D/A range for CO
concentration.
2
D/A range for O
concentration
2
Concentration stability #1.
Concentration stability #2.
Instrument response. Length of each signal processing loop.
Time in parenthesis is standard deviation.
Vacuum pressure.
Sample pressure.
Sample flow rate.
Raw PMT reading.
PMT reading normalized for temperature, pressure, auto-zero
offset, but not range.
UV lamp reading.
UV lamp stability reading.
UV lamp ratio of current reading divided by calibrated reading.
Stray light offset.
PMT dark offset.
UV lamp dark offset.
Slope for current range, computed during zero/span calibration.
Offset for current range, computed during zero/span calibration.
CO
slope, computed during zero/span calibration.
2
CO
offset, computed during zero/span calibration.
2
O
slope, computed during zero/span calibration.
2
O
offset, computed during zero/span calibration.
2
High voltage power supply output.
Reaction cell temperature control duty cycle.
Reaction cell temperature.
O
sensor cell temperature.
2
Internal chassis temperature.
A-23