Remote Operation & Advanced Features
IMPORTANT
168
SAMPLE FLOW WARNING
PMT DET WARNING
UV LAMP WARNING
HVPS WARNING
DARK CAL WARNING
RCELL TEMP WARNING
IZS TEMP WARNING
PMT TEMP WARNING
INVALID CONC
OPERATIONAL FLAGS
In Manual Calibration Mode
In Zero Calibration Mode
In Span Calibration Mode
UNITS OF MEASURE FLAGS
UGM
MGM
PPB
PPM
SPARE/UNUSED BITS
UNASSIGNED FLAGS
Box Temp Warning
Sample Press Warning
System Reset
Rear Board Not Detected
Relay Board Warning
It is possible to assign more than one flag to the same Hessen status bit. This allows the
grouping of similar flags, such as all temperature warnings, under the same status bit.
Be careful not to assign conflicting flags to the same bit as each status bit will be
triggered if any of the assigned flags are active.
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
100. 8000
MP Calibration
Analog Cal Warning
Cannot Dyn Zero
Cannot Dyn Span
Instrument Off
0001
0002
0002
0004
0008
0010
0020
0040
0080
0200
0400
0800
0000
2000
4000
6000