3.5.6. PNEUMATIC LAYOUT WITH CO
Particulate
SAMPLE
GAS INLET
Filter
EXHAUST
GAS OUTLET
PUMP

3.5.7. CALIBRATION GASES

3.5.7.1. ZERO AIR
3.5.7.2. SPAN GAS
Figure 3-22 shows the internal, pneumatic connections for the analyzer with the carbon
dioxide (CO
) sensor option installed. Pneumatically, the CO
2
with the sample gas line between the particulate filter and the analyzer's sample
chamber. It does not alter the gas flow rate of the sample through the analyzer.
CO
2
Probe
Figure 3-22:
Pneumatic Layout with CO
Zero air and span gas are required for accurate calibration.
A gas that is similar in chemical composition to the earth's atmosphere but without the
gas being measured by the analyzer, in this case SO
an Internal Zero Span (IZS) or an external zero air scrubber option, it is capable of
creating zero air.
For analyzers without an IZS or external zero air scrubber option, a zero air generator
such as the Teledyne Model 701 can be used (Figure 3-16).
A gas specifically mixed to match the chemical composition of the type of gas being
measured at near full scale of the desired measurement range. In this case, SO
SENSOR OPTION
2
HYDROCARBON
SCRUBBER
(Kicker)
FLOW
SENSOR
SAMPLE
PRESSURE
FLOW PRESSURE
SENSOR
SENSOR PCA
Sensor
2
Getting Started
sensor is placed in line
2
Chassis
with CO
Sensor Option
2
PMT
REACTION
CELL
. If your analyzer is equipped with
2
UV
LAMP
2
65