Teledyne ML - T100 UV Fluorescence SO2 Analyzer
QUESTION
How often do I need to change the
particulate filter?
What is the averaging time for an
T100?
My analyzer has the optional, user -
configurable analog output channels.
How do I program and use them?
How long does the sample pump last?
Do I need a strip chart recorder or
external data logger?

10.9. TECHNICAL ASSISTANCE

If this manual and its trouble-shooting / repair sections do not solve your problems, technical assistance may be
obtained from Teledyne Monitor Labs, Inc., Customer Service, 35 Inverness Drive East, Englewood, CO 80112.
Phone: 1-800-846-6062. Fax: 1-303-799-4853. Email:
Before you contact customer service, fill out the problem report form in Appendix C, which is also available online
for electronic submission at
User Notes:
10%. Section 8.3.6 includes detailed instructions on performing a check of the
sample gas flow.
Once per week. Table 8-1 contains a maintenance schedule listing the most
important, regular maintenance tasks.
The default averaging time, optimized for ambient pollution monitoring, is 240
seconds for stable concentrations and 20 seconds for rapidly changing
concentrations; Refer to 9.7.1 for more information.
Instructions for this can be found in the Manual Addendum for Configurable Analog
Output, PN 06270.
The sample pump should last about one year and the pump diaphragms should to
be replaced annually or when necessary. Use the PRES test function displayed via
the front panel to see if the diaphragm needs replacement (refer to Section 10.1.2).
No, the T100 is equipped with a very powerful internal data acquisition system.
Section 5.2 describes the setup and operation in detail.
http://www.teledyne-ml.com/manuals.asp
06807A(DCNxxxx)
ANSWER
tml_support@teledyne.com
Troubleshooting & Repair
301