Calibration Procedures

6.4. MANUAL CALIBRATION WITH ZERO/SPAN VALVES

Zero air and span gas are supplied to the analyzer through the zero gas and span gas inlets
rather than through the sample inlet.
The zero and cal operations are initiated directly and independently with dedicated buttons
(CALZ and CALS)
MODEL 700
Gas Dilution Calibrator
(with O
generator option)
3
MODEL 701
Zero Air
Calibrated
Generator
SO
Gas
2
(At high
concentration)
Figure 6-5: Setup for Manual Calibration with Z/S Valve Option Installed (Step 1)
178
Zero and Span calibrations using the Zero/Span Valve option are similar to that
described in Section 7.2, except that:
STEP ONE: Connect the sources of zero air and span gas to the respective ports on the
rear panel (refer to Figure 3-1) as shown below.
Source of
SAMPLE Gas
External Zero
Air Scrubber
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
VENT if input is pressurized
SAMPLE
EXHAUST
SPAN 1
ZERO AIR
Filter
VENT
Needle valve
to control flow
Chassis