O 2 Sensor Operation Within The T100 Analyzer; Figure 9-7: Oxygen Sensor - Theory Of Operation - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Theory Of Operation
9.2.2. O
SENSOR OPERATION WITHIN THE T100 ANALYZER
2
The O
Test functions for O
test functions of the analyzer.
O
sensor calibration is performed via the front panel CAL function and is performed in a
2
nearly identical manner as the standard SO
details.
Stability of the O
232

Figure 9-7: Oxygen Sensor - Theory of Operation

The oxygen sensor option is transparently integrated into the core analyzer operation.
All functions can be viewed or accessed through the front panel display, just like the
functions for SO
.
2
concentration is displayed below the SO
2
slope and offset are viewable from the front panel along with the other
2
sensor can be viewed via the front panel (refer to Section 6.10.1.3).
2
The O
concentration range is 0-100% (user selectable) with 0.1% precision and
2
accuracy.
The temperature of the O
loop and can be viewed on the front panel as test function O
The O
sensor assembly itself does not have any serviceable parts and is enclosed in an
2
insulated canister.
06807A(DCNxxxx)
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
concentration.
2
calibration. Refer to Section 6.10.1 for more
2
sensor is maintained at a constant 50°C by means of a PID
2
TEMP.
2