Appendix A-8: Setting Up Communications - Teledyne T100 Instruction Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML - T100 UV Fluorescence SO2 Analyzer

APPENDIX A-8: Setting up Communications

Setting up the communications for integration with a Teledyne Monitor Labs system.
The following describes the settings that should be entered into the analyzer's COMM sub menus.
Ethernet Communication
The following settings must be entered into the comm. menus in order to achieve Ethernet
communication. Analyzers shipped with systems are setup at the factory.
ID: Analyzer specific. See table below.
INET:
DHCP: OFF
INST IP: site specific
GATEWAY IP: leave as factory default or site specific.
SUBNET MASK: leave as factory default or site specific.
TCP PORT: 3000
ONLINE: ON
BAUD RATE: 115200
Analyzer ID Table
SO
MACHINE ID
100
GAS ID
150
Serial Communication
Teledyne Monitor Labs systems communicate to the analyzers via the PLC using serial communication.
The communication protocol used in the analyzer is the Hessen two protocol. When an analyzer is first
turned on, the Hessen variables do not exist in the non-volatile memory. In order to create these
variables, the Hessen protocol must be turned on in COM 1 and then the power must be cycled.
TRS
NO
2
102
200
160
141
APPENDIX A - Version Specific Software Documentation
CO
CO
x
300
360
130
120
O2
O
/CO
2
2
802
170
2
A-41

Advertisement

Table of Contents
loading

Table of Contents