Teledyne ML - T100 UV Fluorescence SO2 Analyzer

4.8.8. PRESSURE CALIBRATION

A sensor at the exit of the sample chamber continuously measures the pressure of the
sample gas. This data is used to compensate the final SO
changes in atmospheric pressure when the instrument's TPC feature is turned on (refer
to Section 10.7.3) and is stored in the CPU's memory as the test function PRES (also
viewable via the front panel).
Ensure to use a barometer that measures actual barometric pressure.
To cause the analyzer to measure and record a value for PRES, access the Signal I/O
from the DIAG Menu (refer to Figure 4-17), then press:
DIAG
DIAG
PREV NEXT
DIAG PCAL
2
Adjust these values until the
displayed pressure equals the
pressure measured by the
independent pressure meter.
Figure 4-34:
SIGNAL I / O
NEXT
ENTR
Repeat Pressing NEXT until . .
PRESSURE CALIBRATION
ACTUAL PRES :27.20 IN-HG-A
7
.2
0
ENTR accepts the
EXIT ignores the new
DIAG – Pressure Calibration
Operating Instructions
concentration calculation for
2
EXIT
.
ENTR
EXIT
ENTR EXIT
new value
value