The T100 analyzer can be configured for both 100-120 V and 220-240 V at either 50
or 60 Hz.
To avoid damage to your analyzer, ensure that the AC power voltage matches the
voltage indicated on the Analyzer's model identification label (refer to Figure 3-4)
before plugging the T100 into line power.

3.4.2. CONNECTING ANALOG INPUTS (OPTION)

The Analog In connector is used for connecting external voltage signals from other
instrumentation (such as meteorological instruments) and for logging these signals in the
analyzer's internal DAS. The input voltage range for each analog input is 1-10 VDC.
Pin assignments for the Analog In connector are presented in Table 3-5 .
Table 3-5: Analog Input Pin Assignments
PIN
DESCRIPTION
1
Analog input # 1
2
Analog input # 2
3
Analog input # 3
4
Analog input # 4
5
Analog input # 5
6
Analog input # 6
7
Analog input # 7
8
Analog input # 8
GND
Analog input Ground
1
See Section 5.2 for details on setting up the DAS.
CAUTION
GENERAL SAFETY HAZARD

Figure 3-6: Analog In Connector

DAS
PARAMETER
AIN 1
AIN 2
AIN 3
AIN 4
AIN 5
AIN 6
AIN 7
AIN 8
N/A
Getting Started
1
43