10.7.2. SENSOR MODULE REPAIR & CLEANING
IMPORTANT
10.7.2.1. REMOVING AND REINSTALLING THE SENSOR MODULE
Figure 10-6:
Sensor Module Wiring and Pneumatic Fittings
After any repair or service has been performed on the sensor module, the T100 should be
allowed to warm up for 60 minutes.
Always perform a leak check (refer to Section 8.3.6) and calibrate the analyzer (refer to
Section 6) before placing it back in service.
Several of the procedures in this section either require the sensor module to be removed
from the instrument or are easier to perform if it has been removed.
To remove the Sensor Module:
06807A(DCNxxxx)
Troubleshooting & Repair
285