3.5.1. BASIC CONNECTIONS

Calibrated
SO
Gas
at
2
span gas
concentration
MODEL T701
Zero Gas
Generator
Figure 3-16:
58
Refer to Figure 3-4 and Table 3-2 while making the pneumatic connections as follows:
Connect ¼" gas line not more than 2 m long, from sample gas
SAMPLE inlet
source to this inlet.
When no zero/span/shutoff valve options, also connect line from
calibration gas source to this inlet, but only when a calibration
operation is actually being performed.
Connect exhaust line made of PTEF tubing; minimum O.D ¼", to
EXHAUST outlet
this fitting. The exhaust line should be no longer than 10 meters,
and should lead outside the shelter or immediate area surrounding
the instrument.
Figure 3-16 and Figure 3-17 illustrate pneumatic connections for two of the possible
basic configurations.
VENT
Pneumatic Connections–Basic Configuration–Using Bottled Span Gas