Teledyne T100 Instruction Manual page 183

Uv fluorescence so2 analyzer
Hide thumbs Also See for T100:
Table of Contents

Advertisement

Teledyne ML - T100 UV Fluorescence SO2 Analyzer
Scroll to the STABIL test
function.
Wait until STABIL
falls below 0.5
ppb. This may
take several
minutes.
The value of STABIL
may jump
significantly. Wait
until STABIL falls
below 0.5 ppb. This
may take several
minutes.
Figure 6-10:
STEP TWO: Perform the zero/span check.
SAMPLE
RANGE = 500.000 PPB
< TST TST >
CAL
CALZ
SAMPLE
STABIL=XXX.X PPB
< TST TST > CAL
CALZ
ZERO CAL M
STABIL=XXX.X PPB
< TST TST >
ZERO
SAMPLE
STABIL=XXX.X PPB
< TST TST > CAL
CALZ
SPAN CAL M
STABIL=XXX.X PPB
< TST TST >
ZERO SPAN
Setup for Manual Calibration Check with Z/S Valve or IZS Option (Step 2)
SO2 =XXX.X
CALS
SETUP
SO2 =XXX.X
CALS
SETUP
SO2 =XXX.X
CONC
EXIT
SO2 =XXX.X
CALS
SETUP
SO2 =XXX.X
CONC
EXIT
Calibration Procedures
ACTION:
Record the
SO
readings
2
presented in the
upper right corner of
the display.
ACTION:
Record the
SO
readings
2
presented in the
upper right corner of
the display.
EXIT returns to the main
SAMPLE display
183