Theory Of Operation
9.3.2. CO

OPERATION WITHIN THE T100 ANALYZER

2
The CO
Test functions for CO
other test functions of the analyzer.
CO
2
nearly identical manner as the standard SO
Stability of the CO
9.3.3. ELECTRONIC OPERATION OF THE CO
234
The CO2 sensor option is transparently integrated into the core analyzer operation. All
functions can be viewed or accessed through the front panel display, just like the
functions for SO
.
2
concentration is displayed below the SO
2
slope and offset are viewable from the front panel along with the
2
sensor calibration is performed via the front panel CAL function and is performed in a
sensor can be viewed via the front panel (refer to Section 6.10.2.3).
2
Refer to Section 3.7.3 for information on calibrating the CO
The CO
PCA, which is mounted to the rear side of the Relay Board Mounting Bracket,
2
sensor. It converts the sensor's digital output to an analog voltage that
controls the CO
2
is measured with the motherboard and draws 12 VDC from the analyzer via the relay
card from which converts to fit the power needs of the probe and its own onboard logic.
It outputs a 0-5 VDC analog signal to the analyzer's CPU via the motherboard that
corresponds to the concentration of CO
J12
Purple wire
22 awg
Pin 7
Pin 8
Orange wire
22 awg
Relay PCA

Figure 9-9: CO2 Sensor Option PCA Layout and Electronic Connections

06807A(DCNxxxx)
Teledyne ML - T100 UV Fluorescence SO2 Analyzer
concentration.
2
calibration.
2
SENSOR
2
measured by the probe.
2
LED V8
Serial I/O
(Not Used)
CPU
OVDC 5VDC
Analog
Output
Black wire
Grey wire
22 awg
22 awg
P110
Motherboard
.
2
LED V9
To CO
2
Probe