Map the four user-configurable Analog Outputs to any of a wide variety of "Signals"
present in the instrument and customize their respective configurations.
Refer to Figure 2-52 for the following:
Signal Out: select a Signal for the output.
Min/Max: edit minimum and maximum values associated with the selected Signal.
Calibration Type:
AUTO for group calibration (Figure 2-53) of the analog outputs (cannot be
selected when Current is selected for the Range)
MANUAL for individual calibration (Figure 2-54) of analog outputs where manual
adjustments can be made (the only calibration type allowed when Current is
selected for the Range). See Sections 2.5.8.1 and 2.5.8.2 .
Range: assign a voltage or select Current (refer to Table 2-13).
Recorder Offset: add a zero offset for recording slightly negative readings from noise
around the zero point.
Allow Overrange: check to allow a ± 5% over-range; uncheck to disable over-range if
the recording device is sensitive to excess voltage or current.
74
Teledyne API T200 NO/NO
O
NALOG
UTPUTS

Figure 2-52. Analog Output Configuration Example

/NO
Analyzer with NumaView™ Software