Calibration Procedures
5.1.2. Calibration Gas Traceability
5.1.3. Data Recording Devices
5.2. MANUAL CALIBRATION WITHOUT ZERO/SPAN VALVES
Figure 5-1:
170
All equipment used to produce calibration gases should be verified against
standards of the National Institute for Standards and Technology (NIST). To
ensure NIST traceability, we recommend acquiring cylinders of working gas that
are certified to be traceable to NIST Standard Reference Materials (SRM). These
are available from a variety of commercial sources.
A strip chart recorder, data acquisition system or digital data acquisition system
should be used to record data from the T360's serial or analog outputs. If analog
readings are used, the response of the recording system should be checked against
a NIST traceable voltage source or meter. Data recording device should be
capable of bi-polar operation so that negative readings can be recorded. For
electronic data recording, the T360 provides an internal data acquisition system
(DAS), which is described in detail in Section 4.7.
This is the basic method for manually calibrating the Model T360 CO
without functioning zero/span valve options.
described in the GETTING STARTED (Section 3) of this manual and is repeated
her for you convenience.
STEP ONE: Connect the Sources of Zero Air and Span Gas as shown below.
Pneumatic Connections–Basic Configuration–Using Bottled Span Gas
Teledyne API Model T360/T360M Operation Manual
It is identical to the method
Analyzer
2