Figure 3-28: So 2 Span Gas Setting - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Getting Started
Only appears if either
the optional O
or CO
2
sensors are installed.
The SO
span concentration value
2
is automatically default to
400.0 PPB.
If this is not the the concentration of
the span gas being used, toggle
these buttons to set the correct
concentration of the SO
calibration
2
gas.
72
S
SO
S
G
C
ET
PAN
AS
2
Set the expected SO
span gas concentration. This should be 80% of the concentration
2
range for which the analyzer's analog output range is set.
SAMPLE
RANGE=500.0 PPB
<TST
TST>
CAL
SAMPLE
CO
O2
2
SAMPLE
RANGE TO CAL:LOW
LOW
HIGH
M-P CAL
RANGE=400.0 PPB
<TST TST> ZERO SPAN
M-P CAL
SO2 SPAN CONC:400.0 Conc
0
4
0
Figure 3-28:
Teledyne API - T100 UV Fluorescence SO
ONCENTRATION
SO2= XXXX
SETUP
GAS TO CAL:CO
ENTR
ENTR
SO2= XXXX
CONC
0
.0
0
ENTR EXIT
SO
Span Gas Setting
2
EXIT
Only appears if either
the analyzer is set for
EXIT
DUAL range mode.
EXIT
EXIT ignores the new
setting and returns to
the previous display.
ENTR accepts the new
setting and returns to
the
CONCENTRATION
MENU.
06807F DCN7335
Analyzer
2