Calibration Procedures; Calibration Preparations; Required Equipment, Supplies, And Expendables - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

9.

CALIBRATION PROCEDURES

I
MPORTANT
9.1.

CALIBRATION PREPARATIONS

I
MPORTANT
9.1.1.

REQUIRED EQUIPMENT, SUPPLIES, AND EXPENDABLES

06807F DCN7335
This section describes the calibration procedures for the T100. All of the methods
described in this section can be initiated and controlled through the COM ports.
If you are using the T100 for US-EPA controlled monitoring, refer to
Section 0 for information on the EPA calibration protocol.
The calibration procedures in this section assume that the analog range and units of
measure, range mode, and reporting range have already been selected for the analyzer. If
this has not been done, please do so before continuing (refer to Sections 3.4.4.1 and 5 for
instructions).
It is recommended that the LAMP CAL routine (refer to Section 5.9.6) be
performed prior to all calibration operations.
instrument to account for minor changes due to aging of the UV lamp.
Calibration of the T100 analyzer requires specific equipment and supplies. These
include, but are not limited to, the following:
Zero-air source
Sulfur dioxide span gas source
Gas lines - all gas line materials should be Teflon-type or glass.
A recording device such as a strip-chart recorder and/or data logger (optional).
Traceability Standards
I
R
MPACT ON
EADINGS OR
I
R
MPACT ON
EADINGS OR
D
ATA
D
ATA
This will allow the
179

Hide quick links:

Advertisement

Table of Contents
loading

This manual is also suitable for:

T108uT100hT108T100u

Table of Contents