Troubleshooting & Service
Configuration
Figure 11-4:
11.6.2.

DC POWeR SUPPLY

234
J2: Power
Jumper
Location of Relay Board Power Configuration Jumper
AC Configuration of the pump is accomplished via an in-line, hard wired, set of
connections. Call Teledyne API's Technical Support Department for more information.
If you have determined that the analyzer's AC main power is working, but the unit is
still not operating properly, there may be a problem with one of the instrument's
switching power supplies, which convert AC power to 5 and ±15 V (PS1) as well as +12
V DC power (PS2). The supplies can either have DC output at all or a noisy output
(fluctuating).
To assist tracing DC Power Supply problems, the wiring used to connect the various
printed circuit assemblies and DC powered components and the associated test points on
the relay board follow a standard color-coding scheme as defined in Table 11-4.

Table 11-4: DC Power Test Point and Wiring Color Code

NAME
TEST POINT#
DGND
1
+5V
2
AGND
3
+15V
4
-15V
5
+12V
6
+12R
7
A voltmeter should be used to verify that the DC voltages are correct as listed in Table
11-4. An oscilloscope, in AC mode and with band limiting turned on, can be used to
evaluate if the supplies are excessively noisy (>100 mV peak-to-peak).
Teledyne API - T100 UV Fluorescence SO
COLOR
DEFINITION
Black
Digital ground
Red
Green
Analog ground
Blue
Yellow
Purple
Orange
12 V return (ground) line
Analyzer
2
06807F DCN7335