Technical Assistance; Table 11-4. Temperature Controller - Secondary Parameter Settings - Teledyne T102 Operation Manual

Total reduced sulfur analyzer with thermal converter
Table of Contents

Advertisement

Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
Table 11-4. Temperature Controller – Secondary Parameter Settings
PARAMETER
DISPLAY
NAME
Input type
Digital Filter
Lower input range
Upper input range
High Alarm Type
Low Alarm Type
Alarm Hysteresis
Decimal Point
Process variable
Offset
Set variable
Offset
Celsius/Fahrenheit
Fuzzy Logic
Ignore any other parameters that may appear.

11.7. Technical Assistance

07267B DCN6485
SET TO
0
OFF
(Refer to
COMMENTS,
this row)
5
40
limit
1100
limit
10
10
3
0
0
0
C
ON
If this addendum and its trouble-shooting / repair sections do not solve your problems,
technical assistance may be obtained from TAPI, Technical Support, 9970 Carroll
Canyon Road, San Diego, CA 92131. Phone: +1 858 657 9800 or 1-800 324 5190. Fax:
+1 858 657 9816. Email: sda_techsupport@teledyne.com.
Before you contact Technical Support, fill out the problem report form in Appendix C,
which is also available online for electronic submission at http://www.teledyne-
api.com/forms/index.asp.
COMMENTS
Sets the controller to reverse action (hotter
temperatures = less heating; lower temperatures =
more heating) and turns off the heater should the
thermocouple input fail.
You may have one of three thermocouple Types; the
wire colors distinguish which one you have:
red & black (Type S): set to 6
red & yellow (Type K): set to 3
red & orange (Type N): set to 12
If replacing the thermocouple, carefully follow the
instructions provided in Section 11.5.6.
Sets the response time of the digital noise filter to 5
sec.
Sets the lower end of the controller full scale bandwidth
to 40˚C
Sets the upper end of the controller full scale
bandwidth to 1100˚C
Sets the High alarm point to trigger when the process
value exceeds the high alarm set point only after the
process value rises above the lower limit for the first
time after power up
Sets the Low alarm point to trigger when the process
value exceeds the high alarm set point only after the
process value rises above the lower limit for the first
time after power up
Sets the hysteresis bandwidth for both alarm points to
3% of full scale.
Sets the decimal point for PV & SV readings to NONE
OFF
OFF
Sets controller to perform function on Celsius
temperature scale.
Turns of the controllers advanced fuzz logic feature
which further reduces overshoot and fluctuations.
TROUBLESHOOTING and REPAIR