T102 OPERATING INSTRUCTIONS
6.4. Remote Operation of the Analyzer
6.4.1. Control Inputs
INPUT
A
External Zero Cal
B
External Span Cal
C, D, E, & F
Digital Ground
DC Power For
U
Input Pull Ups
Internal +5v
+
Figure 6-2. Control Inputs with Local 5 V Power Supply
40
The following table and figures supersede the Control Input Pin Assignment table and
the figures for Control Inputs with Local 5 V Power Supply and Control Input with
External 5 V Power Supply in the Control Inputs section in the T101 Manual - PN
07266 respectively.
Table 6-4. T102 Control Input Pin Assignments
STATUS
Zero calibration mode is activated. The mode field of the
display will read ZERO CAL R.
Span calibration mode is activated. The mode field of the
display will read SPAN CAL R.
Unused
Provided to ground an external device (e.g., recorder).
Input for +5 VDC required to activate inputs A - F. This voltage
can be taken from an external source or from the "+" pin.
Internal source of +5V which can be used to activate inputs
Supply
when connected to pin U.
CONTROL IN
A
B
C
D
Teledyne API - T102/501 TRS, Addendum to T101 Operation Manual
CONDITION WHEN ENABLED
E
F
U
+
07267B DCN6485