Changing The O 3 Dryer Particulate Filter - Teledyne T200 User Manual

No/no2/nox analyzer with numaview software
Hide thumbs Also See for T200:
Table of Contents

Advertisement

3. Carefully remove the retaining ring, PTFE o-ring, glass window and filter
4. Replace the filter, being careful that the element is fully seated and
5. Reinstall the PTFE o-ring with the notches up; the glass window, then
6. Close the panel and restart the analyzer.
C
HANGING THE
The air for the O
at its inlet, which prevents dust from entering the ozone dryer and degrading the dryer's
performance over time. Change the filter as follows:
1. Before starting the procedure, check and record the average Rx Cell Press and
the O3 Flow values in the Dashboard. (If not visible, see Section 2.5.3 to
configure).
2. Turn off the analyzer, unplug the power cord and remove the cover.
3. Unscrew the nut around the port of the filter using two 5/8" wrenches.
ATTENTION
4. Take off the old filter element and replace it with a suitable equivalent (Teledyne
API P/N FL-3).
114
Teledyne API T200 NO/NO
element.
centered in the bottom of the holder.
screw on the retaining ring and hand tighten. Inspect the seal between the
edge of filter and the o-ring to assure a proper seal.
O
D
P
RYER
ARTICULATE
3
generator passes through a dryer equipped with a small particulate filter
3
COULD DAMAGE INSTRUMENT AND VOID WARRANTY
Ensure to use proper wrenches.
Hold the main dryer fitting with a 5/8" wrench to ensure that it
does not turn against the dryer.
Performing this procedure improperly or with incorrect tools
creates the risk of causing a significant leak.
/NO
Analyzer with NumaView™ Software
2
x
F
ILTER
083730200A DCN7962