Maintenance Procedures; Changing The Sample Particulate Filter; Figure 10-1: Sample Particulate Filter Assembly - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Instrument Maintenance
10.3.

MAINTENANCE PROCEDURES

10.3.1.

CHANGING THE SAMPLE PARTICULATE FILTER

I
MPORTANT
212
The following procedures need to be performed regularly as part of the standard
maintenance of the T100.
The particulate filter should be inspected often for signs of plugging or excess dirt. It
should be replaced according to the service interval in Table 10-1 even without obvious
signs of dirt. Filters with 1 and 5 µm pore size can clog up while retaining a clean look.
We recommend handling the filter and the wetted surfaces of the filter housing with
gloves and tweezers.
Do not touch any part of the housing, filter element, PTFE retaining ring,
glass cover and the O-ring with bare hands, as contamination can
negatively impact accuracy of readings.
To change the filter according to the service interval in Table 10-1:
1. Turn OFF the analyzer to prevent drawing debris into the sample line.
2. Open the analyzer's hinged front panel and unscrew the knurled retaining ring of the
filter assembly.
Figure 10-1:
Teledyne API - T100 UV Fluorescence SO
I
R
MPACT ON
EADINGS OR
Sample Particulate Filter Assembly
2
D
ATA
06807F DCN7335
Analyzer