Table 9-4: Example Auto-Cal Sequence - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

The CALIBRATE attribute must always be set to OFF for analyzers used in
Note
US EPA controlled applications that have IZS option installed.
Calibration of instruments used in US EPA related applications should
only be performed using external sources of zero air and span gas with an
accuracy traceable to EPA or NIST standards and supplied through the
analyzer's sample port (refer to Section 9.2).
194
The following example sets Sequence 2 to carry out a zero-span calibration every other
day starting at 01:00 on June 4, 2015, lasting 15 minutes. This sequence will start 0.5
hours later each day that it is run.

Table 9-4: Example Auto-Cal Sequence

MODE / ATTRIBUTE
SEQUENCE
MODE
TIMER ENABLE
STARTING DATE
STARTING TIME
DELTA DAYS
DELTA TIME
DURATION
CALIBRATE
VALUE
2
Define Sequence #2
ZERO-SPAN
Select Zero and Span Mode
ON
Enable the timer
04-JUN-15
Start after June 4, 2015
01:00
First Span starts at 01:00
2
Run Sequence #2 every other day
00:30
Run Sequence #2 0.5 h later each scheduled day
15.0
Operate Span valve for 15 min
ON
The instrument will re-set the slope and offset
values for the SO
sequence
COMMENT
channel at the end of the AutoCal
2
06807F DCN7335