Manual Adjustment Of Current Range Analog Outputs; Figure 2-48. Setup For Checking / Calibration Current Output Signal Levels - Teledyne T400 User Manual

Photometric ozone analyzer with numaview software
Table of Contents

Advertisement

2.5.8.2 MANUAL ADJUSTMENT OF CURRENT RANGE ANALOG OUTPUTS

These instructions assume that the Current Loop Option is installed (Section 2.3.1.3).
This option places circuitry in series with the output of the D-to-A converter on the
motherboard that changes the normal DC voltage output to a 0-20 milliamp signal.
Adjusting the signal zero and span levels of the current loop output is done by raising or
lowering the voltage output of the D-to-A converter circuitry on the analyzer's
motherboard. This raises or lowers the signal level produced by the current loop option
circuitry.
The software allows this adjustment to be made in 100, 10 or 1 count increments. Since
the exact amount by which the current signal is changed per D-to-A count varies from
output-to-output and instrument–to–instrument, you will need to measure the change in
the signal levels with a separate, current meter placed in series with the output circuit.
See Figure 2-4 for pin assignments and diagram of the analog output connector.

Figure 2-48. Setup for Checking / Calibration Current Output Signal Levels

An alternate method for measuring the output of the Current Loop converter is to connect
a 250 ohm ±1% resistor across the current loop output in lieu of the current meter (see
Figure 2-4 for pin assignments and diagram of the analog output connector). This allows
the use of a voltmeter connected across the resistor to measure converter output as VDC
or mVDC.
74
Do not exceed 60 V peak voltage between current loop outputs
and instrument ground.
mADC
IN
OUT
I OUT +
I OUT -
ANALYZER
Teledyne API – Model T400 O
CAUTION!
Current
Meter
I IN +
I IN -
Recording
Device
Analyzer
3
083730400A DCN8130