Manual Calibration With Zero/Span Valves; Figure 9-5: Setup For Manual Calibration With Z/S Valve Option Installed (Step 1) - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

9.4.

MANUAL CALIBRATION WITH ZERO/SPAN VALVES

MODEL 700
Gas Dilution Calibrator
(with O
generator option)
3
MODEL 701
Zero Air
Calibrated
Generator
SO
Gas
2
(At high
concentration)

Figure 9-5: Setup for Manual Calibration with Z/S Valve Option Installed (Step 1)

186
Zero and Span calibrations using the Zero/Span Valve option are similar to that
described in Section 9.2, except that:
Zero air and span gas are supplied to the analyzer through the zero gas and span gas
inlets rather than through the sample inlet.
The zero and cal operations are initiated directly and independently with dedicated
buttons (CALZ and CALS)
STEP ONE: Connect the sources of zero air and span gas to the respective ports on the
rear panel (refer to Figure 3-4) as shown below.
Source of
SAMPLE Gas
External Zero
Air Scrubber
VENT if input is pressurized
SAMPLE
EXHAUST
SPAN 1
ZERO AIR
Filter
VENT
Needle valve
to control flow
Chassis
06807F DCN7335