Getting Started
3.3.2.4. PNEUMATIC LAYOUT FOR PRESSURIZED SPAN/AMBIENT ZERO OPTION
Figure 3-20:
58
•
remotely by using the external digital control inputs (refer to Section 8.1.2 and
Section 9.7.1)
•
remotely through the RS-232/485 serial I/O ports (refer to Appendix A-6 for the
appropriate commands)
Sources of zero and span gas must be capable of supplying at least 1.55 L/min.
(maximum 2.5L/min). Both supply lines should be vented outside of the analyzer's
enclosure. In order to prevent back-diffusion and pressure effects, these vent lines
should be between 2 and 10 meters in length.
Figure 3-21 shows the internal, pneumatic connections for the analyzer with the
pressurized span ambient zero option installed.
Pneumatic Layout with Pressurized Span/Ambient Zero Option
Table 3-11 describes the state of each valve during the analyzer's various operational
modes.
Table 3-11: Zero/Span and Sample/Cal Valve Operating States
MODE
VALVE
Sample/Cal
SAMPLE
Zero/Span
Sample/Cal
ZERO CAL
Zero/Span
Sample/Cal
SPAN CAL
Zero/Span
Press Span
Teledyne API - T100 UV Fluorescence SO
CONDITION
Open to SAMPLE inlet
Open to ZERO AIR inlet
Open to zero/span inlet
Open to ZERO AIR inlet
Open to zero/span inlet
Open to SPAN GAS inlet
Open to SPAN GAS inlet
Analyzer
2
06807F DCN7335