5.6.2 R
EBUILDING THE
The diaphragm in the sample pump periodically wears out and must be replaced. A
sample rebuild kit is available – see label on the pump itself for the part number of the
pump rebuild kit. Instructions and diagrams are included with the kit.
Always perform a flow check (Section 5.6.6) and a leak check (Section 5.6.5) after
rebuilding the Sample Pump.
5.6.3 R
EPLACING THE
1. Power OFF the analyzer.
2. Remove the cover from the analyzer.
3. Disconnect the white nylon ¼"-1/8" fitting from the Zero Air Scrubber (See Figure 5-7).
4. Remove the old scrubber by disconnecting the 9/16" fitting at the top of the O
tower, then removing the scrubber.
5. Install the new scrubber by reversing these instructions.