O 2 Sensor Operation Within The T100 Analyzer; Figure 12-7: Oxygen Sensor - Principles Of Operation - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Principles of Operation
12.2.2.
O
SENSOR OPERATION WITHIN THE T100 ANALYZER
2
274
Figure 12-7:
The oxygen sensor option is transparently integrated into the core analyzer operation.
All functions can be viewed or accessed through the front panel display, just like the
functions for SO
.
2
The O
concentration is displayed below the SO
2
Test functions for O
the other test functions of the analyzer.
O
sensor calibration is performed via the front panel CAL function and is
2
performed in a nearly identical manner as the standard SO
Section 9.10.1 for more details.
Stability of the O
2
9.10.1.3).
The O
concentration range is 0-100% (user selectable) with 0.1% precision and
2
accuracy.
The temperature of the O
loop and can be viewed on the front panel as test function O
The O
sensor assembly itself does not have any serviceable parts and is enclosed in an
2
insulated canister.
Teledyne API - T100 UV Fluorescence SO
Oxygen Sensor - Principles of Operation
slope and offset are viewable from the front panel along with
2
sensor can be viewed via the front panel (refer to Section
sensor is maintained at a constant 50°C by means of a PID
2
concentration.
2
calibration. Refer to
2
TEMP.
2
06807F DCN7335
Analyzer
2