Chapter 12 — Drive Troubleshooting
Code
Cause and Solution
0, c, b,
No error or message assigned. There may be a problem with the single-character display
h, E, or
(SCD). Turn the power off, then on and determine whether all segments on the SCD are lit.
F
If so, you may have a down-level version of your 3580 Tape Drive's firmware or this book.
Refer to the latest version of the firmware.
A
Tape drive hardware problem. The tape drive determined that a problem occurred that
degraded the operation of the tape drive, but it did not restrict continued use. If the problem
persists, replace the tape drive. The drive is usable, though the single-character display
(SCD) continues to indicate an error and the status light flashes amber. The error code may
clear when you cycle power to the tape drive or place it in maintenance mode.
B
No error or message is assigned. See error code 8 in this table.
C
The tape drive needs to be cleaned. The error code clears when you clean the drive or place
it in maintenance mode.
D
No error or message assigned. See Error Code 0 in this table.
...
The message display has lost communication with the tape drive. This message appears on
Line 2 of the message display. The three dots may occasionally display during normal
processing.
LTO-2 SCD Codes
If multiple errors occur, the code with highest priority (lowest number)
displays first. Once corrected, the code with the next highest priority
displays, until none remain.
Code
Cause and Solution
0
No error occurred and no action is required. The power was cycled or diagnostics have
finished with no errors.
1
Cooling problem; the recommended operating temperature was exceeded. Perform one or
more of the following:
Ensure that the cooling fan is rotating and is quiet.
Remove any blockage that prevents air from flowing freely through the tape drive.
Ensure that the operating temperature and airflow is within the specified range.
If the operating temperature is within the specified range and the problem persists, replace
the drive.
Note: The error code clears when you power-off the tape drive or place it in maintenance
mode.
286
User Guide
Note:
The SCD is blank during normal operation.
91090000

Advertisement

Table of Contents
loading
Need help?

Need help?

Do you have a question about the T-Series Spectra T200 and is the answer not in the manual?

Questions and answers

This manual is also suitable for:

T380T680

Table of Contents