Pressure Calibration; Figure 5-33: Diag - Pressure Calibration - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

SETUP Menu
5.9.7.

PRESSURE CALIBRATION

122
A sensor at the exit of the sample chamber continuously measures the pressure of the
sample gas. This data is used to compensate the final SO
changes in atmospheric pressure when the instrument's TPC feature is turned on (refer
to Section 5.8, Table 5-2) and is stored in the CPU's memory as the test function PRES
(also viewable via the front panel).
Ensure to use a barometer that measures actual barometric pressure.
To cause the analyzer to measure and record a value for PRES, access the Signal I/O
from the DIAG Menu, SETUP>MORE>DIAG (refer to Figure 5-16), then press:
DIAG
DIAG
PREV NEXT
DIAG PCAL
2
Adjust these values until the
displayed pressure equals the
pressure measured by the
independent pressure meter.
Figure 5-33:
Teledyne API - T100 UV Fluorescence SO
SIGNAL I / O
NEXT
ENTR
Repeat Pressing NEXT until . .
PRESSURE CALIBRATION
ACTUAL PRES :27.20 IN-HG-A
7
.2
0
ENTR accepts the
EXIT ignores the new
DIAG – Pressure Calibration
concentration calculation for
2
EXIT
.
ENTR
EXIT
ENTR EXIT
new value
value
06807F DCN7335