Figure 5-22: Diag - Analog Output - Auto Cal Or Manual Cal Selection For Channels - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
DIAG
Exit to return to
the main
PREV
sample display
DIAG AIO
< SET
Press SET> to select the analog output channel to
be configured. Then press EDIT to continue
DIAG AIO
< SET SET>
DIAG AIO
Figure 5-22:
06807F DCN7335
Analyzer
2
To select manual output calibration for a particular channel, access the Analog I/O
Configuration from the DIAG Menu (refer to Figure 5-16), then press:
ANALOG I / O CONFIGURATION
NEXT
ENTR
AOUTS CALIBRATED: NO
SET>
CAL
CONC_OUT_2:5V, CAL
EDIT
CONC_OUT_2 RANGE: 5V
SET>
EDIT
DIAG – Analog Output – Auto Cal or Manual Cal Selection for Channels
Now the analog output channels should either be automatically calibrated or they should
be set to manual calibration, which is described next.
DIAG AIO
EXIT
< SET
SET>
DIAG AIO
EXIT
< SET
SET>
DIAG AIO
ON
EXIT
Toggle to OFF to enable manual
cal adjustments to the selected
output channel only.
DIAG AIO
EXIT
OFF
SETUP Menu
CONC_OUT_2 REC OFS: 0 mV
EDIT
CONC_OUT_2 AUTO CAL: ON
EDIT
CONC_OUT_2 AUTO CAL: ON
ENTR EXIT
CONC_OUT_2 AUTO CAL: OFF
ENTR EXIT
ENTR accepts the new setting
and returns to the previous menu.
EXIT ignores the new setting and
returns to the previous menu.
EXIT
EXIT
111