Hessen; Table 6-5: Rs-232 Communication Parameters For Hessen Protocol - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
6.6.2.

HESSEN

6.6.2.1. HESSEN COMM PORT CONFIGURATION
I
MPORTANT
Note
06807F DCN7335
Analyzer
2
The Hessen protocol is a Multidrop protocol, in which several remote instruments are
connected via a common communications channel to a host computer. The remote
instruments are regarded as slaves of the host computer. The remote instruments are
unaware that they are connected to a Multidrop bus and never initiate Hessen protocol
messages. They only respond to commands from the host computer and only when they
receive a command containing their own unique ID number.
The Hessen protocol is designed to accomplish two things: to obtain the status of remote
instruments, including the concentrations of all the gases measured; and to place remote
instruments into zero or span calibration or measure mode. API's implementation
supports both of these principal features.
The Hessen protocol is not well defined; therefore while API's application is completely
compatible with the protocol itself, it may be different from implementations by other
companies.
The following subsections describe the basics for setting up your instrument to operate
over a Hessen Protocol network. For more detailed information as well as a list of host
computer commands and examples of command and response message syntax,
download the Manual Addendum for Hessen Protocol from the Teledyne API's website:
http://www.teledyne-api.com/manuals/index.asp.
Hessen protocol requires the communication parameters of the T100's COMM ports to
be set differently than the standard configuration as shown in the table below.

Table 6-5: RS-232 Communication Parameters for Hessen Protocol

Parameter
Standard
Data Bits
Stop Bits
Parity
Duplex
To change the rest of the COMM port parameters and modes, refer to Section 6.
To change the baud rate of the T100's COMM ports, refer to Section 6.2.2.
Ensure that the communication parameters of the host computer are also
properly set.
The instrument software has a 200 ms latency before it responds to
commands issued by the host computer. This latency should present no
problems, but you should be aware of it and not issue commands to the
instrument too quickly.
Communications Setup and Operation
Hessen
8
7
1
2
None
Even
Full
Half
I
R
MPACT ON
EADINGS OR
D
ATA
141