SETUP Menu
5.9.3.3. MANUAL ANALOG OUTPUT CALIBRATION AND VOLTAGE ADJUSTMENT
I
MPORTANT
112
For highest accuracy, the voltages of the analog outputs can be manually calibrated.
Calibration is done through the instrument software with a voltmeter connected across
the output terminals (refer to Figure 5-23). Adjustments are made using the control
buttons by setting the zero-point first and then the span-point (refer to Table 5-7).
The software allows this adjustment to be made in 100, 10 or 1 count increments.

Table 5-7: Voltage Tolerances for Analog Output Calibration

FULL SCALE
ZERO TOLERANCE
0.1 VDC
±0.0005V
1 VDC
±0.001V
5 VDC
±0.002V
10 VDC
±0.004V
Outputs configured for 0.1V full scale should always be calibrated
manually.
See Table 3-1 for
pin assignments
of Analog Out
connector on the
V OUT +
ANALYZER
Figure 5-23:
Teledyne API - T100 UV Fluorescence SO
SPAN VOLTAGE
900 mV
4500 mV
4500 mV
I
R
MPACT ON
EADINGS OR
V
rear panel
+DC
Gnd
V OUT -
Setup for Calibrating Analog Outputs
SPAN TOLERANCE
90 mV
D
ATA
V IN +
V IN -
Recording
Device
06807F DCN7335
Analyzer
2
±0.001V
±0.001V
±0.003V
±0.006V