Flow Calibration; Figure 5-34: Diag - Flow Calibration - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
5.9.8.

FLOW CALIBRATION

06807F DCN7335
Analyzer
2
The flow calibration allows the user to adjust the values of the sample flow rates as they
are displayed on the front panel and reported through COM ports to match the actual
flow rate measured at the Sample inlet. This does not change the hardware measurement
of the flow sensors, only the software calculated values.
To carry out this adjustment, connect an external, sufficiently accurate flow meter to the
SAMPLE inlet (refer to Section 10.3.7 for more details).
Once the flow meter is attached and is measuring actual gas flow, access the Signal I/O
from the DIAG Menu, SETUP>MORE>DIAG (refer to Figure 5-16), then press:
DIAG
DIAG
PREV NEXT
DIAG FCAL
0
Adjust these values until the
displayed flow rate equals the
flow rate being measured by the
independent flow meter.
Figure 5-34:
SIGNAL I / O
NEXT
Repeat Pressing NEXT until . .
FLOW CALIBRATION
ACTUAL FLOW: 607 CC / M
6
0
7
DIAG – Flow Calibration
ENTR
EXIT
.
ENTR
EXIT
ENTR EXIT
ENTR accepts the
new value
EXIT ignores the new
value
SETUP Menu
123

Hide quick links:

Advertisement

Table of Contents
loading

This manual is also suitable for:

T108uT100hT108T100u

Table of Contents