Figure 5-4: Setup Rnge - Single Range Mode - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
5.4.3.1. SINGLE RANGE MODE (SNGL)
SAMPLE
RANGE = 500.000 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
PRIMARY SETUP MENU
SETUP X.X
CFG DAS
RNGE
SETUP X.X
MODE
SET UNIT
SETUP X.X
SNGL
IND
AUTO
06807F DCN7335
Analyzer
2
The default range mode for the analyzer is single range, in which all analog
concentration outputs are set to the same reporting range. This reporting range can be set
to any value between 0.1 ppb and 20,000 ppb.
While the two outputs always have the same reporting range, the span and scaling of
their electronic signals may also be configured differently (e.g., A1 = 0-10 V; A2 = 0-
0.1 V).
To select SNGLE range mode and to set the upper limit of the range, press:
SO2 =XXX.X
SETUP
ENTR
EXIT
PASS CLK MORE
EXIT
RANGE CONTROL MENU
EXIT
RANGE MODE:
SNGL
ENTR EXIT
Figure 5-4: SETUP RNGE – Single Range Mode
SETUP X.X
RANGE MODE:
SNGL
SNGL IND
AUTO
SETUP X.X
RANGE CONTROL MENU
MODE
SET
UNIT
SETUP X.X
RANGE:
500.0 Conc
0
0
5
0
0
.0
SETUP X.X
RANGE CONTROL MENU
MODE SET UNIT
SETUP Menu
ENTR
EXIT
EXIT
ENTR
EXIT
EXIT x 2 returns
to the main
SAMPLE display
EXIT
87