Troubleshooting & Service
11.7.2.3. CLEANING THE PMT LENS AND PMT FILTER
I
MPORTANT
Figure 11-9:
248
The sample chamber should only be opened or cleaned on instructions
from the Teledyne API Technical Support Department.
Be careful not to leave thumbprints on the interior of the sample
chamber. The various oils that make up fingerprints fluoresce brightly
under UV light and will significantly affect the accuracy of the
analyzer's SO
measurement).
2
To clean the PMT Lens and filter:
1. Remove the sensor module as described in Section 11.7.2.1.
Hex Screw Between Lens Housing and Sample Chamber
Teledyne API - T100 UV Fluorescence SO
I
R
MPACT ON
EADINGS OR
2
D
ATA
06807F DCN7335
Analyzer