9.10.2.4. CO
ZERO/SPAN CALIBRATION
2
Press CO2: ENTR to
initiate zero point
calibration of the
O2 sensor.
Press CO2: ENTR to
initiate span point
calibration of the
O2 sensor.
The SPAN key now appears
during the transition from
zero to span.
You may see both keys.
If either the ZERO or SPAN
buttons fail to appear see the
chapter on Troubleshooting
for tips.
204
To perform the zero/span calibration procedure:
SAMPLE
< TST
TST >
Toggle TST> button until ...
SAMPLE
< TST TST >
Allow zero gas to enter the sample port
at the rear of the analyzer.
SAMPLE
< TST TST >
SAMPLE
SO2
CO2
M-P CAL
<TST TST>
M-P CAL
<TST TST>
ENTR
Allow span gas to enter the sample port
at the rear of the analyzer.
SAMPLE
< TST TST >
SAMPLE
SO2
CO2
M-P CAL
<TST TST> ZERO
M-P CAL
<TST TST>
M-P CAL
<TST TST> ENTR
Figure 9-20:
RANGE=50.0 PPM
CO2= XX.XX
CAL
SETUP
CO2 STB=X.XX %
CO2=XX.XX
CAL
SETUP
CO2 STB=X.XX %
CO2=XX.XX
CAL
SETUP
GAS TO CAL:CO2
O2
ENTR
EXIT
CO2 STB=X.XX %
CO2=XX.XX
ZERO
CONC
EXIT
CO2 STB=X.XX %
CO2=XX.XX
CONC
EXIT
CO2 STB=X.XX
CO2=XX.XX
CAL
SETUP
GAS TO CAL:CO2
O2
ENTR
EXIT
CO2 STB=X.XX %
CO2=XX.XX
SPAN
CONC
EXIT
CO2 STB=X.XX %
CO2=XX.XX
ENTR
CONC
EXIT
CO2 STB=X.XX %
CO2=XX.XX
CONC
EXIT
CO
Zero/Span Calibration
2
Set the Display to show
the CO2 STB test
function.
This function calculates
the stability of the CO
measurement.
Wait until CO2 STB
falls below 0.01%.
This may take several
minutes.
Press ENTR to changes
the OFFSET & SLOPE
values for the O
2
measurement.
Press EXIT to leave the
calibration unchanged and
return to the previous
menu.
Wait until CO2 STB
falls below 0.01%
This may take several
minutes.
Press ENTR to changes
the OFFSET & SLOPE
values for the CO
2
measurement.
Press EXIT to leave the
calibration unchanged and
return to the previous
menu.
EXIT at this point
returns to the
SAMPLE menu.
06807F DCN7335