9.5.

MANUAL CALIBRATION WITH IZS OPTION

9.6.
MANUAL CALIBRATION CHECKS WITH IZS OR ZERO/SPAN
VALVES
06807F DCN7335
Under the best conditions, the accuracy off the SO
permeation tube is about ±5%. This can be subject to significant amounts of drift as the
tube ages and the amount of SO
sufficient for informal calibration checks, it is not adequate for formal calibrations and is
not approved for use by the US EPA as a calibration source.
Therefore, for formal calibrations of an instrument with an IZS option installed the
following provisions must be followed.
Zero air and span gas must be supplied to the analyzer through the sample gas inlet
as depicted in Figure 9-1 of Section 9.2.
The calibration procedure must be initiated using the CAL button, not the CALZ or
CALS buttons, using the procedure defined in Section 9.2.
Using the CAL button does not activate the zero/span or sample/cal valves of the
IZS option, thus allowing the introduction of zero air and sample gas through the
sample port from more accurate, external sources such as a calibrated bottle of SO
or a Model T700 Dilution Calibrator.
SAMPLE
< TST TST >
Use for formal
calibration
operations.

Figure 9-8: Manual Calibration with IZS Option

Zero and span checks using the zero/span valve or IZS option are similar to that
described in Section 9.3, with the following exceptions:
On units with an IZS option installed, zero air and span gas are supplied to the
analyzer through the zero gas inlet and from ambient air.
On units with a zero/span valve option installed, zero air and span gas are supplied
to the analyzer through the zero gas and span gas inlets from two different sources.
The zero and calibration operations are initiated directly and independently with
dedicated buttons CALZ and CALS.
contained in the tube is depleted. Whereas this may be
2
RANGE = 500.000 PPB
CAL
CALZ
CALS
effusion rate of the IZS option's
2
SO2 =XXX.X
SETUP
Use only for
informal calibration
checks.
2
189