Data Acquisition System (DAS) and APICOM
Use PREV and NEXT to
scroll to the data channel
to be edited.
Set the number of days
between reports (0-365).
Press buttons to set amount of
time between reports, in hours
(HH) and/or minutes (MM)
(max: 23:59). 01:00 sets a
report to be made every hour.
7.2.6.

NUMBER OF RECORDS

160
To define the REPORT PERIOD, follow the instruction shown in Figure 7-3, then
press:
Edit Data Channel Menu
SETUP X.X
0) CONC:
PREV NEXT
SETUP X.X
NAME:CONC
<SET
SET>
EDIT
Press SET> until you reach REPORT PERIOD ...
SETUP X.X
REPORT PERIOD:000:00:05
<SET
SET>
EDIT
SETUP X.X
REPORT PERIODD:DAYS:0
0
0
0
SETUP X.X
REPORT PERIODD:TIME:01:00
0
1
0
IIf at any time an invalidl entry is selected (e.g., days > 366)
the ENTR button will disappear from the display.
Figure 7-8: DAS – Define the Report Period
Although the DAS is capable of capturing several months worth of data,, the actual
number of records is also limited by the total number of parameters and channels and
other settings in the DAS configuration. Every additional data channel, parameter,
number of samples setting etc. will reduce the maximum amount of data points
somewhat. In general, however, the maximum data capacity is divided amongst all
channels (max: 20) and parameters (max: 50 per channel).
The DAS will check the amount of available data space and prevent the user from
specifying too many records at any given point. If, for example, the DAS memory space
can accommodate 375 more data records, the ENTR button will disappear when trying
to specify more than that number of records. This check for memory space may also
make an upload of a DAS configuration with APICOM or a Terminal program fail, if
the combined number of records would be exceeded. In this case, it is suggested to
either try from the front panel what the maximum number of records can be or use trial-
and-error in designing the DAS script or calculate the number of records using the DAS
or APICOM manuals. To set the number of records for one channel from the front panel,
press SETUP-DAS-EDIT-ENTR and the following control button sequence.
Teledyne API - T100 UV Fluorescence SO
ATIMER,
2,
4032, R
INS
DEL
EDIT
PRNT
EXIT
PRINT
EXIT
PRINT
EXIT
ENTR
EXIT
0
ENTR
EXIT
Analyzer
2
Exits to the main
Data Acquisition
menu.
ENTR accepts the new string and
returns to the previous menu.
EXIT ignores the new string and
returns to the previous menu.