Figure 3-10: Control Input Connector; Table 3-8: Control Input Signals - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
CONTROL IN
A
B
C
D
Local Power Connections

Table 3-8: Control Input Signals

Input #
Status Definition
A
REMOTE ZERO CAL
REMOTE SPAN CAL
B
C, D, E & F
SPARE
Digital Ground
U
External Power input
+
5 VDC output
06807F DCN7335
Analyzer
2
+
E
F
U
Figure 3-10:
ON Condition
The analyzer is placed in Zero Calibration mode. The mode field of the
display will read ZERO CAL R.
The analyzer is placed in span calibration mode as part of performing a low
span (midpoint) calibration. The mode field of the display will read LO CAL
R.
The ground level from the analyzer's internal DC power supplies (same as
chassis ground)
Input pin for +5 VDC is required to activate pins A – F.
Internally generated 5V DC power. To activate inputs A – F, place a jumper
between this pin and the "U" pin. The maximum amperage through this port
is 300 mA (combined with the analog output supply, if used).
A
B
C
-
External Power Connections
Control Input Connector