Overview of Operating Modes
Table 4-1: Analyzer Operating Modes
MODE
DIAG
1
LO CAL A
1
LO CAL R
1
M-P CAL
SAMPLE
SAMPLE A
SETUP
2
SPAN CAL A
2
SPAN CAL M
2
SPAN CAL R
2
ZERO CAL A
2
ZERO CAL M
2
ZERO CAL R
1
Other calibration procedures under CAL mode are described separately in Section 9.
2
Only Appears on units with Z/S valve or IZS options.
76
The Mode field of the front panel display indicates to the user which operating mode the
unit is currently running.
In addition to SAMPLE and SETUP, other modes available are presented in Table 4-1.
One of the analyzer's diagnostic modes is active (refer to Section 5.9).
Unit is performing LOW SPAN (midpoint) calibration initiated automatically by the analyzer's
AUTOCAL feature
Unit is performing LOW SPAN (midpoint) calibration initiated remotely through the COM ports or
digital control inputs.
This is the basic calibration mode of the instrument and is activated by pressing the CAL button.
Sampling normally, flashing text indicates adaptive filter is on.
Indicates that unit is in SAMPLE mode and AUTOCAL feature is activated.
SETUP mode is being used to configure the analyzer. The gas measurement will continue during
this process.
Unit is performing SPAN calibration initiated automatically by the analyzer's AUTOCAL feature
Unit is performing SPAN calibration initiated manually by the user.
Unit is performing SPAN calibration initiated remotely through the COM ports or digital control
inputs.
Unit is performing ZERO calibration procedure initiated automatically by the AUTOCAL feature
Unit is performing ZERO calibration procedure initiated manually by the user.
Unit is performing ZERO calibration procedure initiated remotely through the COM ports or digital
control inputs.
Teledyne API - T100 UV Fluorescence SO
Figure 4-1: Front Panel Display