Figure 11-8: Sample Chamber Mounting Bracket - Teledyne t100 Operation Manual

Uv fluorescence so2 analyzer
Hide thumbs Also See for t100:
Table of Contents

Advertisement

Teledyne API - T100 UV Fluorescence SO
11.7.2.2. CLEANING THE SAMPLE CHAMBER
I
MPORTANT
06807F DCN7335
Analyzer
2
The sample chamber should only be opened or cleaned on instructions
from the Teledyne API Technical Support department.
Be careful not to leave thumbprints on the interior of the sample
chamber. The various oils that make up fingerprints fluoresce brightly
under UV light and will significantly affect the accuracy of the analyzer's
SO
measurement)
2
To clean the sample chamber:
1. Remove the sensor module as described in Section 11.7.2.1.
2. Remove the sample chamber mounting bracket by unscrewing the four bracket
screws.
Figure 11-8:
3. Unscrew the 4 hexagonal standoffs.
4. Gently remove the chamber cover.
5. Using a lint-free cloth dampened with distilled water, wipe the inside surface of the
chamber and the chamber cover.
6. Dry the chamber surfaces with a 2nd lint-free cloth.
7. Re-assemble the chamber and re-install the sensor module.
I
R
MPACT ON
EADINGS OR
Sample Chamber Mounting Bracket
Troubleshooting & Service
D
ATA
247