SETUP Menu
5.4.3.

REPORTING RANGE MODES: SINGLE, DUAL, AUTO RANGES

SAMPLE
RANGE = 500.000 PPB
< TST TST > CAL
SAMPLE
ENTER SETUP PASS : 818
8
1
8
SETUP X.X
CFG DAS
RNGE
PASS CLK MORE
Only one of the
range modes may be
active at any time.
86
The T100 provides three analog output range modes to choose from:
Single range (SNGL) mode sets a single maximum range for the analog output. If
single range is selected (refer to Section 5.4.3.1) both outputs are slaved together
and will represent the same measurement span (e.g. 0-50 ppm); however, their
electronic signal levels may be configured for different ranges (e.g. 0-10 VDC vs. 0-
.1 VDC – Refer to Section 5.9.3.1).
Dual range (DUAL) allows the A1 and A2 outputs to be configured with different
measurement spans (refer to Section 5.4.3.2).
Auto range (AUTO) mode gives the analyzer to ability to output data via a low
range and high range. When this mode is selected (refer to Section 5.4.3.3) the T100
will automatically switch between the two ranges dynamically as the concentration
value fluctuates.
Also, in this mode the RANGE Test function displayed on the front panel during
SAMPLE mode will be replaced by two separate functions, Range1 and Range2.
Range status is also output via the External Digital I/O Status Bits (refer to Section
8.1.1).
To select the Analog Output Range Type press:
SO2 =XXX.X
SETUP
ENTR
EXIT
EXIT
See the section on
Single Range Mode
Figure 5-3: SETUP RNGE – Reporting Range Mode
Teledyne API - T100 UV Fluorescence SO
SETUP X.X
RANGE CONTROL MENU
MODE
SET UNIT
SETUP X.X
RANGE MODE:
SNGL
SNGL DUAL AUTO
See the section on
Dual Range Mode
2
EXIT
EXIT Returns
to the Main
ENTR EXIT
SAMPLE Display
See the section on
Auto Range Mode
06807F DCN7335
Analyzer